做了一次串联染色,降解严重,找了一下BD的官网,觉得说的不错拷贝过来供大家学习。
总之 Be careful when staining tandem dyes like PE-Cy7 and APC-Cy7, the Cy7 moiety is damaged byfixation.

A Guide to Tandem Dyes and Degradation
In this blog, we'll describe the common characteristics of tandem dyes, discuss their advantages in multicolor flow cytometry, and address a common myth about tandem degradation.

What is a tandem dye?

A tandem dye is composed of two covalently attached fluorescent molecules. One of these molecules is either a protein-based (i.e., PE, APC) or synthetic dye (i.e., BV421™) characterized by a large extinction coefficient (capacity to absorb energy) that serves as a donor, while the second molecule is a small synthetic dye that serves as an acceptor (i.e., Cy5, Cy7).

Why do we use tandem dyes?

As you may already know, the number of fluorophores one is able to use at the same time is limited by the number of lasers and detectors on the cytometer (Table 1). The most advanced cytometer instruments on the current market have 5 lasers including ultraviolet (355 nm), violet (405 nm), blue (488 nm), green (532 nm) or yellow-green (561 nm), and red (633 nm) and the potential to integrate 8 parameters off of each laser. However, there are not currently enough spectrally distinct fluorophores yet to fill out this configuration. Without tandem fluorophores, each laser would only be suitable to discretely excite a small number of fluorophores, thus severely limiting the total number of parameters able to be detected.

                               
登录/注册后可看大图


Table 1. Example of multicolor flow panel diversification based on available lasers, detectors, and instrument.


Further expansion of the number of detectors off of each laser has allowed us to increase flow panels up to 21 colors /23 parameters (Table 1). It is possible to achieve because the tandem dyes use the same excitation characteristics as the donor dye but possess emission properties of the acceptor. Therefore, though donor dyes and their tandems are excited by the same laser, they can be used in the same panel by utilizing different detectors to readout their emission. As an example, Figure 1 demonstrates the excitation and emission spectrum of BV421™ and its tandems.

The further expansion of multicolor panels can be achieved by using a spectral cytometer, such as Cytek's Aurora™ Spectral Cytometer. This instrument has 48 fluorescent channels with three lasers. Theoretically, it can detect 42 fluorophores as long as those fluorophores have distinct emission profiles. A spectral detection system does not utilize traditional methods of compensation calculation and allows a simultaneous use of fluorophores with significant overlapping emission spectrum (i.e. BV750™ and BV785™) while having minimal compensation concerns.

                               
登录/注册后可看大图


Figure 1. Excitation and emission spectra of BV421™ and its tandems, BV570™, BV605™, BV650™, BV711™, BV750™, BV785™.


How do tandem dyes work?

The donor fluorophore absorbs light energy of the specific wavelength. Upon excitation, energy is transferred from the donor to the acceptor through a phenomenon called Förster resonance energy transfer (FRET), also known as fluorescence resonance energy transfer. The acceptor emits the transferred energy as fluorescent light. When FRET efficiency is high, a strong signal will be observed in the acceptor channel and a weaker signal in the donor channel. Please note, the FRET efficiency is never 100%, which means that some spillover or bleeding in the donor channel is expected.

I see a fluorescence signal in the donor channel. Is my tandem degrading?

NO. This is a common myth. The donor and acceptor dyes are covalently conjugated and don't typically fall apart. As mentioned earlier some spillover or bleeding into the donor channel is expected since FRET efficiency is not 100%. To accurately compensate spillover, it is important to use the same antibody used in the multicolor sample as the single color compensation control and to treat them identically as the multicolor sample especially in light and fixative exposure. However, if you observe a strong signal in the donor channel and a weak signal in the acceptor channel, this is an indication of low FRET efficiency that is caused by one of the following factors:
  • Photobleaching. Always protect tandem dyes from light or other sources of oxidative stress as they are highly susceptible to photobleaching resulting from oxidation.
  • Exposure to freezing temperature. Do not freeze tandem dye antibody conjugates as it might result in denaturation of the protein-based donor fluorophore.
  • Fixation and Permeabilization. Don't leave your cells in fixative for long periods of time as it will result in increase of the autofluorescence and cause greater harm to the tandems. It is strongly advised to wash the cells after fixation and replace the buffer with the FluoroFix™ Buffer. You can also perform surface staining after fixation. However, be sure to check our Fixation page for compatibility of our antibody clones with fixation.
As an example of how different fixatives may affect the fluorescence signal of tandem dyes, Figure 2 demonstrates the fluorescence signal from cells stained with a PerCP/Cy5.5-conjugated antibody and exposed to 1%PFA or True-Phos™ Perm buffer. Exposure to the True-Phos™ Perm buffer, which is methanol-based, leads to the denaturation of PerCP since this is a protein based dye. It results in the loss of signal for PerCP but not for its acceptor, Cy5.5 whose fluorescence can be detected in the Alexa Fluor® 700 channel.

                               
登录/注册后可看大图


Figure 2. A) Fluorescence intensity of PerCP/Cy5.5 anti-CD14 antibody staining upon exposure to 1% PFA or True-Phos™ Perm buffer solution. B) Dot plots demonstrating appearance of signal in Alexa Fluor® 700 channel upon exposure of PerCP/Cy5.5 to True-Phos™ Perm buffer.


Hopefully, now you have a better understanding of tandem dyes and are familiar with guidelines on their use and handling when planning and doing your experiments. As always, if you have any questions while planning your experiments, feel free to reach out to our technical services team at tech@biolegend.com. You can learn more about these fluorophores with our tandem dyes webpage.

组织样本处理不好?流式中文网原研的魔滤®魔杵®套装,低成本解决,高质量收获
 楼主| 发表于 2018-12-11 00:35:43 | 显示全部楼层
顺便问一句大家,如果是新抗体新sample跑流式,到底PE-CY7会不会发一点PE的光?因为我标记的抗体比较多,PE刚好买的是标记数量很少的一群细胞,所以一点非特异都会造成影响。
组织样本处理不好?流式中文网原研的魔滤®魔杵®套装,低成本解决,高质量收获
发表于 2018-12-12 08:37:51 | 显示全部楼层
liu_liuxn 发表于 2018-12-11 00:35
顺便问一句大家,如果是新抗体新sample跑流式,到底PE-CY7会不会发一点PE的光?因为我标记的抗体比较多,PE ...

未降解的时候,那个叫荧光渗漏,不叫PE-cy7发一点PE的光。

可通过FMO对照确认有效性。
流式中文网FlowGuard®流式专用保存液,无需冻存,稳定保护各类流式样本,从容完成实验
 楼主| 发表于 2018-12-14 13:08:10 | 显示全部楼层
niwanmao 发表于 2018-12-11 19:37
未降解的时候,那个叫荧光渗漏,不叫PE-cy7发一点PE的光。

可通过FMO对照确认有效性。 ...

谢谢倪老师,等回国了一定参加网站的活动哈哈
组织样本处理不好?流式中文网原研的魔滤®魔杵®套装,低成本解决,高质量收获
发表于 2018-12-14 19:51:21 | 显示全部楼层
liu_liuxn 发表于 2018-12-14 13:08
谢谢倪老师,等回国了一定参加网站的活动哈哈

也期待你能够分享国外的生活和学习工作:)
流式中文网FlowGuard®流式专用保存液,无需冻存,稳定保护各类流式样本,从容完成实验
您需要登录后才可以回帖 登录 | 加入流式中文网

本版积分规则 需要先绑定手机号

手机版|流式中文网 ( 浙ICP备17054466号-2 )

浙公网安备 33038202004217号

GMT+8, 2025-3-15 03:05

Powered by Discuz! X3.5

© 2001-2025 Discuz! Team.

快速回复 返回顶部 返回列表